C COFB 088 048 PRIMING DETECTOR TEST ERROR,
BLOOD IN PRIMING DETECTOR (SECOND TEST), CPU P
Unconditional error, blood pump not available
General description of conditions for occurring:
- CPU P sensed “blood” when the priming detector test intensity is high again
after the low intensity test..
Technical description of conditions for occurring:
- PRI(PI_PrimDetEmptyP2ErrCnt) > 10
Some possible explanations why the error has occurred:
- Priming detector not connected to FM I/O board (P504)
- Software error
- Priming detector is not empty or priming detector dirty
- Priming detector board is broken
- Priming detector transmitter electronics broken (IC55, IC33, Q9, Q13 and
components) on protective driver board.
- Priming detector receiver electronics broken (IC18, V11, V12 and
components) on protective driver board.
C COFB 088 049 PRIMING DETECTOR TEST ERROR,
BLOOD IN PRIMING DETECTOR (SECOND TEST), CPU W |
|