发表于 2012-4-12 00:21:14
|
显示全部楼层
Message:
E
CT_XRS_49
CT_XRS_E_50_31__P1P2
Text:
Control error XRS (%1)
Explanation:
UI_LOAD_BAD error is detected.
Possible cause:
- Tube defective
- Generator adaptation incorrect
- Inverter defective
- HV_TANK defective
- Cable connection not OK
Action:
1. Check connection D400 X8 and X9.
2. Perform Generator Adaptation.
3. Perform HVC01 Test with "Test Plugs".
4. If OK, getter/replace tube.
5. If not OK, Use HVC03 Single Pulse Test.
6. If the Shape is OK, replace the HV_TANK.
7. If the shape is not OK, replace the Electronic box.
Information:
FWtext: UI load
P1: n.a.
P2: n.a.
|
|