[设备故障求助] 狼牌气腹机2232.001报错代码为E004希望能有人能提供解决办法?

  [复制链接]
发表于 2014-2-21 18:08:55 | 显示全部楼层 |阅读模式
狼牌气腹机2232.001报错代码为E004希望能有人能提供解决办法?
发表于 2014-2-23 10:09:17 | 显示全部楼层
Error 01: CO2--cylinder closed or empty
Error 02: Button depressed when switching on
Error 03: Power supply not within tolerance
Error 04: V1 defective ; flow measurement defective; V3 defective
Error 05: V2 defective, Insufflation tube connected
Error 06: Pressure measurement defective
Error 07: Flow measurement defective
Error 08: V3 defective ; V2 defective ; pressure measurement defective; Flow measurement defective
Error 09: nominal value invalid
Error 10: Helium
Error 11: Current pressure  30 mmHg for more than 5 Seconds
Error 12: General hardware error; pressure measurement defective; V1 defective;
flow measurement defective
Error 13: Pressure switch ( input pressure) defective
Error 14: Electronic pressure switch defective
Error 15: Wrong DIL switch coding (electric pressure monitoring fast / slow) or pressure switch defective
回复 支持 1 反对 0

使用道具 举报

发表于 2014-3-3 17:47:36 | 显示全部楼层
故障解决了吗?没有解决可以咨询我,电话是13260315894张或QQ361101490
已绑定手机
发表于 2014-3-27 15:35:53 | 显示全部楼层
2楼的这个是什么牌子的啊才?这个报错代码的话最好能找到说明书查一查,上面一般都有